(a) An instrument is calibrated in an environment at 20C and the following output readings y are obtained for various input values x: y 13.1 26.2 39.3 52.4 65.5 78.6 x 5 10 15 20 25 30 Determine the...



(a) An instrument is calibrated in an environment at 20C and the following output


readings y are obtained for various input values x:


y 13.1 26.2 39.3 52.4 65.5 78.6


x 5 10 15 20 25 30


Determine the measurement sensitivity, expressed as the ratio y/x.


(b) When the instrument is subsequently used in an environment at 50C, the


inputeoutput characteristic changes to the following:


y 14.7 29.4 44.1 58.8 73.5 88.2


x 5 10 15 20 25 30


Determine the new measurement sensitivity. Then determine the sensitivity drift


due to the change in ambient temperature of 30C.